[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]
RE: Improvements to cooling system make it worse
One more thing. The thermometers used, AD590, are rated at +/- 2 C. They
are usually much better than that, but since they put out 1 ua per degree,
(i.e 273 ua at 0 C) one is taking difference of big numbers at our
operating temperature. You will note that I subtract 273 in the
temperature computation. This is all done through a 1% resistor. So even
if the AD590 was perfect, one could read 270 and another 276 at zero C just
from the resistor. Add to this the 2 degree spec of the AD590 and one
could get 268 with one device and 278 with another, or +/- 5 C in absolute
measurement.
The moral of all this is that you have to calibrate the thermometers if you
want to know what the real temperature of everything is. One way to do
this is to let the whole mess settle to ambient temperature. Turn off all
the TEC drives and circulate room temperature water. Turn off the dew
heaters. Now the two CCD sensors, and the water temperature sensor should
all come to the same temperature. Be sure you are not putting heat into
the cooling water with the pump. Now you can put a calibrated thermometer
into the coolant, and put entries into the read out table to make the
readings correct at that one temperature. Now on another day, with
another temperature, one can get another set of readings. This should let
you adjust the slope of each calibration to make all the thermometers track.
As you all can see, this is a big project, and not of much use. In fact,
the temperature readings tend to be much better than the possible errors
indicate. As I say, the emphasis in the design was on holding the
temperature constant as the water temperature changes. This it does with a
significant gain. It should have a gain of at least 100. That is a 1
degree change in the water should change the CCD temperature less than 0.01
C if everything is working. I did serious testing on the Mark III. On the
bench I was able to hold to 0.001 C with a gain of about 1000. I have not
done this for the Mark IV. It is the same general design with the same
amplifiers. So it should be similar except the camera head has more
thermal mass for the Mark IV.
Tom Droege
At 10:04 PM 1/8/02 -0500, you wrote:
>I was trying to test my mount pointing tonight but just when the Pleides
>came into view so did the clouds. It's going to be cloudy for a while so
>I'll provide some comparison numbers for TEC current, temps, etc.
>
>Value TOM Mike
>TEC Power 11.954 11.35
>+15 14.798 14.66
>+5 4.88 4.86
>Ground -0.02 0.00
>-5 -5.048 -5.10
>-15 -15.338 -15.01
>VCO Ref. 5.020 5.03
>ADC Ref 2.514 2.49
>DAC 14 set -2 -2
>DAC 15 set -2 -2
>DAC 14 read -1.992 -1.98
>DAC 15 read -2.036 -2.03
>
>Temp CCD0 -19.58 -23.64
>Curr CCD0 0.875 0.78
>Temp CCD1 -19.37 -23.06
>Curr CCD1 0.813 0.91
>Water Temp 7.27 4.80
>
>Tom's system and mine show almost identical values. With nearly the same
>currents we get nearly the same cooling below ambient, in this case about
>28C below the water temp.
>
>
>Mike G.